A spectrophotometer was calibrated with a 6.768 x 10-4 M (in Cu) Cu-neocuproine solution. This calibration was carried o

Posted: Tue Jul 05, 2022 8:58 am
by answerhappygod
A Spectrophotometer Was Calibrated With A 6 768 X 10 4 M In Cu Cu Neocuproine Solution This Calibration Was Carried O 1
A Spectrophotometer Was Calibrated With A 6 768 X 10 4 M In Cu Cu Neocuproine Solution This Calibration Was Carried O 1 (44.25 KiB) Viewed 14 times
A spectrophotometer was calibrated with a 6.768 x 10-4 M (in Cu) Cu-neocuproine solution. This calibration was carried out in a 1.00 cm cell, and yielded an absorbance of 0.2820 AU. An unknown was then analyzed in this same spectrophotometer, but using a 2.50 cm cell. If the absorbance of the unknown was 0.4880 AU, what is the Cu concentration? 4.0 M