Consider the following common scenario, where an unused installed coaxial cable in a fridge is closed off at the top dur
Posted: Fri May 20, 2022 8:01 pm
Consider the following common scenario, where an unused installed coaxial cable in a fridge is closed off at the top during a cooldown where it is not required: 300K 50K 20 4K The Lat the top indicates that the line has been terminated with a matched 50 Ohm impedance at the room temperature coaxial input port. The 20 indicates that a 20 dB attenuator has been installed at (and thermally clamped to) the 4K fridge stage. Assuming room temperature is at 300 K and the nominal "4K" plate is currently running at 3.81 K, follow the instruction in the next