6) (5 points) In the special topic box in Section 6.7, the extreme sensitivity of scanning tunnelling microscopes (STM)

Posted: Mon May 02, 2022 2:29 pm
by answerhappygod
6 5 Points In The Special Topic Box In Section 6 7 The Extreme Sensitivity Of Scanning Tunnelling Microscopes Stm 1
6 5 Points In The Special Topic Box In Section 6 7 The Extreme Sensitivity Of Scanning Tunnelling Microscopes Stm 1 (26.8 KiB) Viewed 43 times
6) (5 points) In the special topic box in Section 6.7, the extreme sensitivity of scanning tunnelling microscopes (STM) was described. It was reported that a change in the tunnelling gap of only 0.4 nm between STM sample and probe can change the tunnelling current by a factor of 104. Check the plausibility of this statement by using the "wide-barrier" approximation in Equation (6.70) to find the difference between the electron energy and barrier height that would produce such a situation. Report your answer in eV and comment in the result